Bjarne,
Thanks for your reply. I hope to be trying out 1 or 2 suggestions to see
if we can make any difference...
To answer your questions-
(1) Yes the tail is dependant on the signal level the poor CTE is only
really seen at low signal levels. We pre-flash the chip with some
LEDs we have in the dewar to give us a few hundred units signal. The
trail in the H overscan is what we have left after 50+1024
(underscan+image) horizontal shifts. At higher signal levels the pixel
level falls back to the bias within a pixel or 2 into the over-scan.
(2) Yes we had a problem with the oSW causing feed-thru and only finally
overcame its effects (hi bias, noise) by performing a transition down
and up between the 2 halves of the CDS. I agree performing the oSW with
oH clocks negates the use of this phase for summing and, as important,
we make the oSW transitions fairly rapid to keep the time between the
CDS REF and SAMP as short as possible.
Cheers
Paddy
PS,
anyone want to buy a TEK1024- we had one sent as replacement for an out-of-spec
device from SITe. They had measured the device parameters at -45, not -90 as
their glossy states and so they were willing to change it. Apparently SITe now
charge 10% EXTRA for a spec. at -90deg C !!!!!
We will have to send it back as the chip we got is okay and has been sent to
La Palma, well, we still have the CTE problem mentioned on my Web Pages....
--
___________________ * _-_
\==============_=_/ ____.---'---`---.____ *
\_ \ \----._________.----/
* \ \ / / `-_-' *
* __,--`.`-'..'-_
/____ || *
`--.____,-' ...to boldly go where no one has gone before!
REFORM:: 'To train, develop or mould by instruction, discipline
or example...'
Captain Paddy...
--
More information about the CCD-world
mailing list