SITe QE values ??
Paul Jorden
prj at ast.cam.ac.uk
Thu Jun 13 11:44:44 CLT 1996
John, & others,
> We are looking to buy a goodly number of large CCDs in the very
> near future. One of the stumbling blocks may be our QE desires,
> summarized below:
> Recently, SITe has informed us that they may perhaps be able to
> comply with the above requirements. My question:
> Does anyone out there have actual QE measurements of thinned SITe
> devices that approach the above. I certainly have never seen anything
> approaching these values, but perhaps someone has some unpublished
> measurements.
> Comments would be welcomed. JG.
I've compiled my table, with John's desires, and three chips measured at RGO at
cryogenic temperatures. I've added some extra wavelength points also.
Thr RGO-measured chips are- a recent EEV CCD42 sample chip, a 5-yr old standard
TK1024, a recent Lesser-thinned 2k3eb Loral chip (2-layer AR).
WAVELENGTH JG desire RGO-QE (EEV42) RGO-QE (TK1024) RGO-Lesser
330 50% 32 19 72
350 38 26 73
400 85 65 54 71
500 85 63 80
600 85 86 69 88
700 85 73 69 90
800 75 51 56 81
900 50 25 29 55
1000 5 7 21
Firstly no chips have matched all of JG's desires.
As we know only the Lesser process has demonstrated the exceptional UV figures;
but even his current 2-layer coatings do not seem able to provide 85% at 400nm.
I suspect that a coating that gave 85% at 400nm would fall off more into the
red, and not provide the 800/900nm desired response. Any comments Mike??
It is also fairly clear that the chip cannot be made too thin, if it is to have
a high red response. In this case, many chips with normal resistivity silicon
(20-100 ohm) are not fully depleted, and will suffer poor blue MTF. This is seen
in standard Lesser-thinned chips. I think you know this John, but I'm
broadcasting this message also..
I recall considering the Gemini CCD specs, and I think they had the same
problem- ie looking for high UV and far-red response, as well as high peak
figures. My guess is Mike Lesser could get closest. However, a commercial
supplier (like SITe or EEV) could offer better quality devices reliably, but
surely would struggle with the 330nm response; and if they managed to achieve
it, I don't think they could offer good 900nm response at present!
I've seen Alistair's recent figures; my comment is that the SITe UV figures look
pretty good, but I'd always prefer to know cryogenic values.
Good luck. I'll be interested in any comments from others.
Regards, Paul
-------------------------------------------------------------------
E-mail: prj at ast.cam.ac.uk Machine: Kria.ast.cam.ac.uk
Royal Greenwich Observatory, Madingley Road, Cambridge, CB3 0EZ, UK.
Phone: +44 (0) 1223- 374000 \direct phone- 374811 \RGO Fax- 374700
More information about the CCD-world
mailing list