Grease
Paul Jorden
prj at ast.cam.ac.uk
Mon Feb 12 11:25:50 CLST 1996
Andrew,
To continue the grease discussion:
> I was rivetted by you statement that:
>
>> Finally, In case you don't know, we have now learnt that silicone 'high vac.
>> grease' is bad news in a cryostat. It is clear that it can become volatile
>> and deposits itself on surfaces. (It's vapour pressure was quoted as ~ 10-6
>> mbar; this is clearly not low enough!) We no longer use it. We use only a
>> small trace of 10-12 mbar grease (non-silicone) now on o-rings only.
>
> I have been using Apiezon L, M and N with impunity (thermal coupling
> O-rings, etc.) I have, of late, had contamination problems with
> my thin chips.
>
> Tell me more ....
>
We used to use Dow Corning High Vacuum grease; this is silicone, with a quoted
vapour pressure of < 10-7 mbar I believe.
We used it as a thin layer to improve thermal contact under our Tek/SITe 1K CCDs
and the cold block.
We definitely got contamination within the cryostat from this source- it's a
long story, but mass spectrometer analysis confirmed it. It is 'well known' that
this vacuum grease is actually bad in this respect!
Apiezon 'M' (a mineral-based hydrocarbon grease) is quoted as having a vapour
pressure of < 10-7 mbar also.
We have now switched to using Fluorinated Fomblin grease FM090 from Balzers;
this has a vapour pressure < 10-12 mbar. It costs us #200 per 50 gm, but we
don't use much of it! (Only the minimum on 'o' rings now, never elsewhere..)
For improved thermal contact we now use a thin layer of indium where necessary.
Cheers, Paul Jorden
-------------------------------------------------------------------
E-mail: prj at ast.cam.ac.uk Machine: Kria.ast.cam.ac.uk
Royal Greenwich Observatory, Madingley Road, Cambridge, CB3 0EZ, UK.