CCD pixel-to-pixel variations

Klaus Reif reif at astro.uni-bonn.de
Wed Mar 17 12:21:52 CLST 1999


Posted to CCD-world:
-+-+-+-
Hello Scott,

>QUESTION #1:
>	Since no light is falling on the CCD during the dark frame
>	exposure, what value of the PRNU would I use, since it
>	shows such dramatic variations as a function of wavelength
>	in the table above?
>

   we have found that the pixel-to-pixel QE variation (i.e. PRNU) 
and the pixel-to-pixel variation of thermal electron generation
(no acronym so far?) are different by an order of magnitude or
more. We have seen values for the latter of 20 to 30 percent
(standard deviation) at room temperature. PRNU values were 
usually between a few and less then 1 percent as indicated 
in the EEV table. These two effects should therefore be considered 
and corrected separately. 

Klaus Reif


*******************************************************************