Contamination
Paul.Jorden at eev.com
Paul.Jorden at eev.com
Mon Jan 18 13:37:07 CLST 1999
Posted to CCD-world:
-+-+-+-
Chris, & CCD-worlders,
Hi. This problem sounds familiar!
I refer you to a 1996 discussion, which you can find at Tim Abbotts'
ccd-world' archive.
go to http://www.cfht.hawaii.edu/~tmca/CCD-world/ and look at 1996.
For your convenience I repeat a message that I posted then; in the days of
RGO!!!
Hope it helps.
Best Regards,
Paul
---------------------------
Re: Grease
Paul Jorden (prj at ast.cam.ac.uk)
Mon, 12 Feb 1996 10:25:50 GMT
Messages sorted by: [ date ][ thread ][ subject ][ author ]
Next message: Tim Abbott: "Re: getters/molecular (fwd)"
Previous message: Paul Jorden: "Closed Cycle coolers?"
Andrew,
To continue the grease discussion:
> I was rivetted by you statement that:
>
>> Finally, In case you don't know, we have now learnt that silicone 'high
vac.
>> grease' is bad news in a cryostat. It is clear that it can become
volatile
>> and deposits itself on surfaces. (It's vapour pressure was quoted as ~
10-6
>> mbar; this is clearly not low enough!) We no longer use it. We use only a
>> small trace of 10-12 mbar grease (non-silicone) now on o-rings only.
>
> I have been using Apiezon L, M and N with impunity (thermal coupling
> O-rings, etc.) I have, of late, had contamination problems with
> my thin chips.
>
> Tell me more ....
>
We used to use Dow Corning High Vacuum grease; this is silicone, with a
quoted
vapour pressure of < 10-7 mbar I believe.
We used it as a thin layer to improve thermal contact under our Tek/SITe 1K
CCDs
and the cold block.
We definitely got contamination within the cryostat from this source- it's a
long story, but mass spectrometer analysis confirmed it. It is 'well known'
that
this vacuum grease is actually bad in this respect!
Apiezon 'M' (a mineral-based hydrocarbon grease) is quoted as having a
vapour
pressure of < 10-7 mbar also.
We have now switched to using Fluorinated Fomblin grease FM090 from Balzers;
this has a vapour pressure < 10-12 mbar. It costs us #200 per 50 gm, but we
don't use much of it! (Only the minimum on 'o' rings now, never elsewhere..)
For improved thermal contact we now use a thin layer of indium where
necessary.
Cheers, Paul Jorden
_______________________________________________________________
Dr Paul Jorden, CCD Applications, EEV
Waterhouse Lane, Chelmsford, Essex, CM1 2QU, UK
Tel: 44 (0) 1245 -453458 (direct), -493493 (switchboard)
Fax: 44 (0) 1245 453224
see also:- http://www.ccd.eev.com/
-+-+-+-
For information about CCD-world, send email to owner-CCD-world at cfht.hawaii.edu.