CCD-world: Reset gate threshold shift
GregJ888 at aol.com
GregJ888 at aol.com
Tue Apr 25 14:40:00 CLT 2000
The following was posted to CCD-world:
ESD is surprisingly misunderstood. Many of the continuity monitors work by
placing a constant current source on the operator and will peak at or above
18v DC. To check -- measure from the sense portion of the wrist strap to
ground with a volt meter (or unground the wrist strap but leave the monitor
grounded). Also check the operator when strapped, some units don't alarm
until above 12v. Shunt the strap to ground with a variable resister. Find
the set point where the alarm sounds and measure with you volt meter.
Some other things we've found (measurements below are dV/dT field
measurements):
1) "ESD dissipative" freeze spray is capable of generating a 2000v field when
sprayed-- as are will most spray cleaners. Tribo-charging of the liquid
moving though the nylon spray tube. I keep ALL sprays away from static
sensitive devices.
2) Charge/Cap= Vconstant right? If the Charge is on the human body, then by
lifting 1 foot (Cap/2), V about doubles.
3) Silk tie, 2000v. Boss's ESD smock has to be zipped all the way up.
4) Charge separation, i.e. getting out of a chair-- 1000v easy, 2000v if
humidity is low. So stand up , then take the wrist strap off.
5) A broken hand held ionizer will put out +100v.
FYI- I have a business relationship with Novx so be advised. Their monitors
are use in the disk drive industry with a 5v ESD limit settable to 1v. Sorry
about the plug, further info- <A HREF="http://www.novxcorp.com/">Novx</A>
Check out the 53XX, that's what's on my bench when a cryostat is open.
Greg Jones
In a message dated 4/24/00 1:25:26 PM Pacific Daylight Time,
roger at ctios1.ctio.noao.edu writes:
<<
What could have happened? Static discharge is rarely a problem in our
very humid environment. Nonetheless, the CCD has always been handled
on properly grounded antistatic mats (bench and floor) while using a
wrist strap with a continuity alarm. Remember too that I obtained
normal images with normal voltages immediately prior to taking the CCD
to Cerro Tololo and it was kept under vacuum and cold the whoel time
until it ceased to operate.
>>
- -- CCD-world -- --
CCD-world is fully moderated. Send posts to CCD-world at astro.ku.dk
Standard replies will go to the list; address personal replies manually.
For more information, please go to: http://www.not.iac.es/CCD-world/
More information about the CCD-world
mailing list